I'm in the market for a used spectrum analyzer like the 7L12
or the HP 141-T. One important use will be to make two-tone IMD measurements on HF SSB transmitters. I am concerned that with tone separations on the order of one kHz, the 7L12 may not have sufficient bandwidth in the 300 Hz mode to resolve adjacent IMD products expected to differ in amplitude by 40 dB or more. Has anyone on the group used a 7L12 for this purpose and is the 300 Hz RBW sufficiently narrow? There seems little doubt that the 141, with 10 Hz or 100 Hz RBW, will handle this. Many thanks in advance. Chuck NT3G |
#2
![]() |
|||
|
|||
![]()
On Wed, 18 May 2005 19:04:44 -0400, chuck wrote:
I'm in the market for a used spectrum analyzer like the 7L12 or the HP 141-T. One important use will be to make two-tone IMD measurements on HF SSB transmitters. I am concerned that with tone separations on the order of one kHz, the 7L12 may not have sufficient bandwidth in the 300 Hz mode to resolve adjacent IMD products expected to differ in amplitude by 40 dB or more. Has anyone on the group used a 7L12 for this purpose and is the 300 Hz RBW sufficiently narrow? There seems little doubt that the 141, with 10 Hz or 100 Hz RBW, will handle this. Many thanks in advance. Chuck NT3G I have never played with a 7L12 but do have a 141t. 300hz bandwidth will work. The best resolution on the 141t is 100hz. There is a 10 hz video filter but the video filter does no good for resolution. It only gets rid of noise. The biggest problem that I can imagine with the 7L12 would be whether or not your scope has some kind of storage or long persistency on the tube. To look at 300 hz bandwidth you need a very slow sweep speed on the scan. If there is no storage you will not be able to see it. 73 Gary K4FMX |
#3
![]() |
|||
|
|||
![]()
Thanks for responding, Gary.
Sweep time is something that I had not even been considering, so I appreciate the info. From what I can gather about the 7L12, the slowest sweep is 10 mSec/division, or 0.1 second for a complete sweep. That is the sweep setting Tek says to use for all frequency domain analyses, regardless of bandwidth. The other sweep settings go from 10 mSec/division to something like 1 uSec/division, but they are intended for time domain analyses. As a reality check on a 0.1 second sweep, I looked at HP's performance test procedure for the 8552B to see what sweeps they specify for measuring the 8552B's bandwidth. For the 300 Hz RBW, they list a 0.2 second sweep (I assume that's not 0.2 second/division). Not really too far from Tek's 0.1 second. Of course, at the narrower bandwidths, HP's specified sweeps get much slower, as you pointed out. And if it is 0.2 second/division, then there is a profound difference between the Tek and HP filter designs or I'm missing something big. As I recall, the slower sweeps are to avoid ringing in the filters, gaussian skirts notwithstanding. So maybe the 7L12 doesn't require a storage scope for the 300 Hz RBW? Sure wish I had access to the Tek Op manual for the 7L12. 73, Chuck Gary Schafer wrote: On Wed, 18 May 2005 19:04:44 -0400, chuck wrote: I'm in the market for a used spectrum analyzer like the 7L12 or the HP 141-T. One important use will be to make two-tone IMD measurements on HF SSB transmitters. I am concerned that with tone separations on the order of one kHz, the 7L12 may not have sufficient bandwidth in the 300 Hz mode to resolve adjacent IMD products expected to differ in amplitude by 40 dB or more. Has anyone on the group used a 7L12 for this purpose and is the 300 Hz RBW sufficiently narrow? There seems little doubt that the 141, with 10 Hz or 100 Hz RBW, will handle this. Many thanks in advance. Chuck NT3G I have never played with a 7L12 but do have a 141t. 300hz bandwidth will work. The best resolution on the 141t is 100hz. There is a 10 hz video filter but the video filter does no good for resolution. It only gets rid of noise. The biggest problem that I can imagine with the 7L12 would be whether or not your scope has some kind of storage or long persistency on the tube. To look at 300 hz bandwidth you need a very slow sweep speed on the scan. If there is no storage you will not be able to see it. 73 Gary K4FMX |
#4
![]() |
|||
|
|||
![]()
chuck wrote:
Thanks for responding, Gary. Sweep time is something that I had not even been considering, so I appreciate the info. From what I can gather about the 7L12, the slowest sweep is 10 mSec/division, or 0.1 second for a complete sweep. That is the sweep setting Tek says to use for all frequency domain analyses, regardless of bandwidth. The other sweep settings go from 10 mSec/division to something like 1 uSec/division, but they are intended for time domain analyses. As a reality check on a 0.1 second sweep, I looked at HP's performance test procedure for the 8552B to see what sweeps they specify for measuring the 8552B's bandwidth. For the 300 Hz RBW, they list a 0.2 second sweep (I assume that's not 0.2 second/division). Not really too far from Tek's 0.1 second. Of course, at the narrower bandwidths, HP's specified sweeps get much slower, as you pointed out. And if it is 0.2 second/division, then there is a profound difference between the Tek and HP filter designs or I'm missing something big. As I recall, the slower sweeps are to avoid ringing in the filters, gaussian skirts notwithstanding. So maybe the 7L12 doesn't require a storage scope for the 300 Hz RBW? Sure wish I had access to the Tek Op manual for the 7L12. 73, Chuck Hi Chuck, I have used both the 141T system, and the 7L13. I ditched my 141T because the 7L5, 7L13, and 7L18 plugins perform better than the equivalent HP plugins for the 141T system, and are much more compact. But, When I was looking into 7L analyzers, I discounted the 7L12. It is too primative. The minimum you want to do any real work is a 7L13 with a 7633 storage frame. The 7L14 is much better because it has the digital storage, but it is also twice the price of a good 7L13. You *will* need a storage scope frame for either the 12, or the 13. The narrow bandwidth sweeps must be done really slowly. If you try and rush them, you will lose most of the amplitude information... the filters just cannot respond quickly. It is a physical reality of narrow band filters. -Chuck Harris |
#5
![]() |
|||
|
|||
![]()
From one Chuck to another,
Thanks for the info and the advice! I will definitely look further into the '13 and '14. I do understand the need to sweep "slowly" at narrow RBWs. But I'm still troubled by the fact that the *slowest* sweep built into the 7L12 is 10 msec/division! That will, arguendo, degrade the filter response. The storage scope will surely not sweep the SA at a slower rate, and putting a distorted SA output signal into a storage scope can't possibly reshape the response! So there is no cure. If our assumptions are correct, this is a fatal Tek design flaw (not a whole lot of them around). A storage scope would be really important if the SA is sweeping too slowly for the regular scope's persistence, or to capture a single-sweep trace. Or for simply storing a trace for later viewing. But if the sweep rate is 10 sweeps/second, there shouldn't be much flicker with P31. Something is amiss here, I think. Maybe there is a typo in Tek's spec sheet? Or more likely, a parity bit error in my cpu! Chuck |
#6
![]() |
|||
|
|||
![]()
chuck wrote:
From one Chuck to another, Thanks for the info and the advice! I will definitely look further into the '13 and '14. I do understand the need to sweep "slowly" at narrow RBWs. But I'm still troubled by the fact that the *slowest* sweep built into the 7L12 is 10 msec/division! That will, arguendo, degrade the filter response. The storage scope will surely not sweep the SA at a slower rate, and putting a distorted SA output signal into a storage scope can't possibly reshape the response! So there is no cure. If our assumptions are correct, this is a fatal Tek design flaw (not a whole lot of them around). A storage scope would be really important if the SA is sweeping too slowly for the regular scope's persistence, or to capture a single-sweep trace. Or for simply storing a trace for later viewing. But if the sweep rate is 10 sweeps/second, there shouldn't be much flicker with P31. Something is amiss here, I think. Maybe there is a typo in Tek's spec sheet? Or more likely, a parity bit error in my cpu! Chuck No, your data sheet is wrong. The slowest automatic sweep is 10 secs per division. The slowest sweep is manual. This specification exists across the entire 7L line. The 7L5, 7L13, 7L14, and 7L18 all have monitors built in that will show "uncal" if you sweep too fast for the filter setting. -Chuck |
#7
![]() |
|||
|
|||
![]()
On Thu, 19 May 2005 20:15:16 -0400, chuck wrote:
From one Chuck to another, Thanks for the info and the advice! I will definitely look further into the '13 and '14. I do understand the need to sweep "slowly" at narrow RBWs. But I'm still troubled by the fact that the *slowest* sweep built into the 7L12 is 10 msec/division! That will, arguendo, degrade the filter response. The storage scope will surely not sweep the SA at a slower rate, and putting a distorted SA output signal into a storage scope can't possibly reshape the response! So there is no cure. If our assumptions are correct, this is a fatal Tek design flaw (not a whole lot of them around). A storage scope would be really important if the SA is sweeping too slowly for the regular scope's persistence, or to capture a single-sweep trace. Or for simply storing a trace for later viewing. But if the sweep rate is 10 sweeps/second, there shouldn't be much flicker with P31. Something is amiss here, I think. Maybe there is a typo in Tek's spec sheet? Or more likely, a parity bit error in my cpu! Chuck I just looked at the 141t and it takes 2ms/div for no flicker. This is at wide bandwidth. At 2khz/div spectrum width and 300 hz bandwidth it takes 50ms/div sweep speed max. At 5 khz spectrum width it takes .1 sec/div sweep speed. At 100hz bandwidth and 2khz/div spectrum width it takes .5sec/div sweep speed. Definitely storage scope area! The 50ms/div sweep can be viewed without storage on a p31 but it is not good. 73 Gary K4FMX |
#8
![]() |
|||
|
|||
![]()
Thanks for the additional info, Gary.
As you can see, I'm having a difficult time understanding just what the 7L12 does. There may be a typo in the data sheet, but when I look at the time/div dial on the 7L12, it just doesn't have the numbers to do anything slower than 10 ms/div. The other models in the 7L series have many more sweep speeds available, based both on their data sheets and on photos. I suspect that the 7L12 was different. In fact, because the 7L13 has 100 Hz and 30 Hz RBWs, it would have to have slower sweep rates. Your finding with the 141 was that with a 300 Hz bandwidth and 2 kHz/div width, a sweep rate of 50 ms/div was marginal. I imagine then that a sweep rate of 10 ms/div (assuming that would be compatible with the 300 Hz filter) would be ok with P31. If I understand correctly. I think we've gone about as far as we can without either having a 7L12 in front of us or at least an op manual. Thanks again. 73, Chuck |
#9
![]() |
|||
|
|||
![]()
On Thu, 19 May 2005 20:49:29 -0400, Chuck Harris
wrote: chuck wrote: From one Chuck to another, Thanks for the info and the advice! I will definitely look further into the '13 and '14. I do understand the need to sweep "slowly" at narrow RBWs. But I'm still troubled by the fact that the *slowest* sweep built into the 7L12 is 10 msec/division! That will, arguendo, degrade the filter response. The storage scope will surely not sweep the SA at a slower rate, and putting a distorted SA output signal into a storage scope can't possibly reshape the response! So there is no cure. If our assumptions are correct, this is a fatal Tek design flaw (not a whole lot of them around). A storage scope would be really important if the SA is sweeping too slowly for the regular scope's persistence, or to capture a single-sweep trace. Or for simply storing a trace for later viewing. But if the sweep rate is 10 sweeps/second, there shouldn't be much flicker with P31. Something is amiss here, I think. Maybe there is a typo in Tek's spec sheet? Or more likely, a parity bit error in my cpu! Chuck No, your data sheet is wrong. The slowest automatic sweep is 10 secs per division. The slowest sweep is manual. This specification exists across the entire 7L line. That's not what this says: http://www.tucker.com/images/images_spec/00000453.pdf I wouldn't assume that the resolution BW of the 7L12 would be adequate to look at IMD down 40+ dB at 1 KHz spacing. The BW is specified as 300 Hz at -6dB with a -6 to -60 dB shape factor of 4:1. I don't have time to plot the selectivity curve at the moment, but it might be an exercise for you. If you have a "perfect" cw signal as input, when you sweep it, what you are plotting on the screen is the filter response of the SA. Figure a Gaussian response with the specified shape factor and then overlap two of the curves with 1 KHz spacing and see if the filter skirts are down 50 dB where they overlap. (You should have 10 dB of margin IMHO) Personally I'd use the HP with a storage mainframe. |
#10
![]() |
|||
|
|||
![]() No, your data sheet is wrong. The slowest automatic sweep is 10 secs per division. The slowest sweep is manual. This specification exists across the entire 7L line. That's not what this says: http://www.tucker.com/images/images_spec/00000453.pdf I pulled out a 1981 Tek catalog, and I am indeed remembering wrong. This is yet another reason why I discounted the 7L12 as a credible SA. Here is what Tek is expecting you to do. There are timebase positions for 5ms through 0.01us. These are for use when the SA is being used as a receiver (time domain mode), and you are looking at a received pulse train. The "SA" mode is *variable* from 10ms through 5ms/division. You are expected to manually adjust the sweep rate to get a clear picture. So, how can you get 300Hz filter resolution? Well, simply by not scanning the full bandwidth of the SA plugin. Reduce the sweept bandwidth to 10KHz, and 300Hz is easily achieved with a 5-10ms/division sweep rate. Not a great way to go, but usually if you are interested in the narrow resolutions, you are only looking for signals over a narrow bandwidth. As I said earlier, get a 7L13, or 7L14. The 7L12 wasn't fully incubated when it was hatched. When you go looking at the 141T, remember, it is a mid 1960's SA design, and it feels like it when you use it. The 7L5, 7L13, and up were designed in the very late 1970s, and take advantage of things like microprocessors to help with house keeping operations. They are smaller, quite reliable, and just plain work better than the 141T family. (And, yes I have owned, used and repaired both.) The 141T storage tube is a nightmare. Very short life. Another SA line that is usually very inexpensive, and much better than the 141T family, is the Eaton/Ailtech 727 and up. I was told (in the early '80s) by an HP FAE (who specialized in HP's SA's and other RF gear) that the Ailtechs were all over the place in HP's internal R&D labs. They were the SA's that HP used in designing their own product line. (flame suit on, helmet latched, as I await the onslot of rebuttles from HP guys..) -Chuck Harris |